ERROR 07LH OPTIMUS 50 PHILIPS ANY IDEA WHO IS PROBLEM
thank`s
measure the tube coil ,it's about 10 ohm
and check contactor like k2/k3
may be the rotor control board breakdown
I have a problem with my Optimus 50 .
When I switch on the system generator :
- The self tests start :
In the starting phase all status LEDS on the Z rack flash . A red illuminated LED indicates
, a self test run .
On the control desk C all displays illuminated temporarily
- At the end of self test :
The red status LED lights off : system is ready , but the relay ENK1 is not switches on and there is no indicates in the desk C ( KV; mas , s ) ? please help me
Quote Nbribakeur
I have a problem with my Optimus 50 .
When I switch on the system generator :
- The self tests start :
In the starting phase all status LEDS on the Z rack flash . A red illuminated LED indicates
, a self test run .
On the control desk C all displays illuminated temporarily
- At the end of self test :
The red status LED lights off : system is ready , but the relay ENK1 is not switches on and there is no indicates in the desk C ( KV; mas , s ) ? please help me
ENK1 will come in if EN100K2 come in which will be on if EZ130 (kV ctrl) is OK. So, check: 1.are all 3 phases ON, 2.is neutral N correct, 3.does red blinking LED turns off after initialisation at EZ130?-if stays it is deffect 4.does PSU correct? (+5, +/-15 and +26V) 5.finally, if all these before OK it is very possible EZ139 (CU, brain). Good luck! available at Viber +38763893095 Ekrem.medpoint@bih.net.ba
07LH is related to missmach of measured and expected anode rotation speed. You can check to test does it make difference to put tube axis in horizontal and vertical position. Can listen bearing sound (make PREP and go near tube to hear possible bearing sound)! It could be machanical bearing problem; old tube?
ekrem.medpoint@bih.net.ba, viber +38763893095
Hello felow engineers,
Ima working on philips optimus 50 generator.
The system is rady but no exposure,
hand swicth ok
please help.