WBC Chamber Error

This error appeared so  i checked the tubings sorrounding the WBC chamber. They are without any clog so i should check the Filter. Whereabout is the WBC chamber filter??

 

Michael


   RE: Hematology Analyzer Report Post
Mon Jun 16 2014Reply from Eduardo Mc Grath
avatar placemark

Ceck the line between the WBC Chamber and SV 10. Change the filter N° 13 Assy C-2, please use the new version of filter you can find with the following part number 023-1962-7. Also see inside of the WBC Chamber, the also version chamber some times release the coating Teflon.


Report Post
Mon Jun 16 2014Reply from Michael Aike
avatar placemark

Thanks alot Mc Grath.

Is washing, cleaning and changing the filter requires taking out the whole chamber?

 

Michael


Report Post
Mon Jun 16 2014Reply from Eduardo Mc Grath
avatar placemark

I mean that the old version WBC Chamber with coating Teflon. The new version had new treatment process and you identify by a black dot in the upper side. For change the filter is not necessary to taking out the WBC Chamber, but be carfully to disconect the tubing under WBC Chamber and pull the tubing gentle for change the filter. Warning take all the liquid remain before remove the filter with a syringe with out using needles.


Report Post