Wed Jun 18 2014

E24 Error Code

When I power the system up I get an error E24 and I have noticed that the collimator is not calibrating to a home position. No collimator movement at all.

F3 is ok

Measured ribbon cable through U45.

What to try?


   RE: Mobile C-Arm Report Post
Wed Jun 18 2014Reply from Karim Fouad
avatar placemark

Check the 15v dc, if ok check u 45.


Report Post
Sun Jun 22 2014Reply from Atlanta Imaging Services
avatar placemark

E 24 has to do with the collimator pot. You either have a bad swivel harness or 325 board. You have to ohm out the cable wires to/from collimator, there should be a couple spare wires in the harness.

Glen


Report Post