Pressure 0.06MPa shows 0.0000 sometimes 0.0582 (fluctuate reading)
-0.03MPa shows -0.000 to -0.0306 also (fluctuate reading).
I ran the correction for Air Pump, nothing solved.
Should I also check tubing air-leaks
Michael Aike
Pressure 0.06MPa shows 0.0000 sometimes 0.0582 (fluctuate reading)
-0.03MPa shows -0.000 to -0.0306 also (fluctuate reading).
I ran the correction for Air Pump, nothing solved.
Should I also check tubing air-leaks
Michael Aike
Each year you must change the Air Pump. If an old Pump, is 100 % that air Pump start with problem, and also is very important also to change the tubing from the Pinch Valve. Let me know and also send the error reports.
Quote Eduardo Mc Grath
Each year you must change the Air Pump. If an old Pump, is 100 % that air Pump start with problem, and also is very important also to change the tubing from the Pinch Valve. Let me know and also send the error reports.
Hi Eduardo -- can you email me for help with a sysmex system? james(at)senecascientific.com. Thank you.