Pressure Low Error

Pressure 0.06MPa shows 0.0000 sometimes 0.0582 (fluctuate reading)

-0.03MPa shows -0.000 to -0.0306 also (fluctuate reading).

I ran the correction for Air Pump, nothing solved.

Should I also check tubing air-leaks

Michael Aike


   RE: Hematology Analyzer Report Post
Wed Jul 16 2014Reply from Eduardo Mc Grath
avatar placemark

Each year you must change the Air Pump. If an old Pump, is 100 % that air Pump start with problem, and also is very important also to change the tubing from the Pinch Valve. Let me know and also send the error reports.


Report Post
Thu Oct 08 2015Reply from Seneca Scientific
avatar placemark

Quote Eduardo Mc Grath

Each year you must change the Air Pump. If an old Pump, is 100 % that air Pump start with problem, and also is very important also to change the tubing from the Pinch Valve. Let me know and also send the error reports.

 