Thu Jul 24 2014

Siemens acuson x300

Having a problem switching a system on.... Ac and power light come on at the tray. But system won't boot up..no lights on on button on cp console.

Cable from oi module to cp console checked. Tp7 on io=0v.no voltage on any tp's ac tray. I do have 5v on standby test point on dc>doc pcb...

Please help I'm going crazy.........


   RE: Ultrasound Systems Report Post
Mon Sep 28 2015Reply from Biomedical Engineer
avatar placemark

TP7 on I/O module= +12V, supplying power for HDD, CD-R/W througg I/O.

I think you should recheck BE board.

Test point TP7 on DC|DC board= 0V???


Report Post
Wed Sep 30 2015Reply from US Specialist
avatar placemark

what you mean by tp7

DCDC powersupply TP 

if so  i want you to check up the 48v  at ACDC power supply

then check the DC DC power supply TP

did you hear the fans sound its working??

monitor led goes green??

all DC dc TP is ok except 12v  then check between DCdc and BE board

most likly DCdc power supply


Report Post
Thu Nov 26 2020Reply from FITCH
avatar placemark

hi, cab you help me to know value on tp 1-18 en dc-dc board.

 

thanks.


Report Post