when i turn on my HP Hewlet packard sonos 2000 Ultrasound machine a message appears on the screen that contains "An error has occurred contact your HP services representative EE566-29490A00"
An error has occurred
Check your PVPS, power supply board A9. With a linear transducer connected, test point VPP2 should be 40 - 46V, VDD should be 10V +/- .2, VNNB & VNNA will vary from -1 to -109V, PV-INT will be 4 to 6V, and DIS-BIAS will be -82 to -131V. There are two .5A fuses on the board.
Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
Report Post
Hi Sir, Can u please help me to get. S/W and service manual for hewlett packard 2000 sonos, thanks to this Mail. waleedhaden@hotmail.com
waleedhaden10@gmail.com
Report Post
Although it is your choice to go offline with your discussion, we are sure that other MedWrench users would appreciate it if you kept your discussion within the thread. This will help to keep a permanent record of the issue you are facing and how it was solved to help others in the community that may be facing a similar challenge.
We appreciate your participation and hope that the MedWrench forums will help you to solve your medical equipment issue.
Report Post