Start up is halted. Shutdown is recommended
Reboot to restore system functionality
Service ID : US_IMG_24.when I put probe on it. That wht is giving me on the screen .how can I solve this problem
Start up is halted. Shutdown is recommended
Reboot to restore system functionality
Service ID : US_IMG_24.when I put probe on it. That wht is giving me on the screen .how can I solve this problem
Hello - Your problem is either with the Analogue power supply or the RC pcb. When changing either of these parts it is important to run firmware loading and both calibrations.
Service ID: US_IMG_33 or Eh::StatusItem Exception caught errors are poping up when starting the Xario XG machine is because you need to replace with new RAMS on the Motherboard.
My Xario XG started working perfectly after i had done the above solution.
https://www.youtube.com/channel/UC3fzl2x_Lvfwpc3iRNB9nYQ
Dinushan Wickramarathna
Thanks