005 Error code

3.0.2.711 SW, E cart.

IE33 is giving intermittent 005 error code. In checking Tech Admin, it is happening with different transducers.

The error log generally stacks like this in Tech Admin

Error:  Fault Power monitor

Error: Received Power Monitor ADUE

EHA005 Fault Dialog 005.8639.4XPJ.16 displayed

Error Recovery: User reselected scanhead

Error Fault:  Power Monitor

Is this an FEC issue? Help?

Also found Error DIalog 005.BWRU.8YMD.5 displayed


   RE: Ultrasound Systems Report Post
Mon Oct 06 2014Reply from Kyle_Summit
Kyle_Summit

I've Found when the 005.bwru is intermitant its typically a connection problem either lose ribbons to the FEC and DSC in the back end, or Burnt Molex connectors on the AFP, I would suggest visualy inspecting it if you can. 

In the info log it should be tracking what probe was being used during the time it failed, it would also but useful to see if it was the same probe each time. 

feel free to contact us if you have any questions



kyle@mysummitimaging.com

Manager of Global Education & Training

Website: http://www.mysummitimaging.com/

Phone:  1-866-586-3744

24/7