Would anyone have the tech lit for the Philips Veradius?  We're getting a "M395 Battery OK, hardware error in generator" as soon as we try to expose.

The error is the same in the BV family manual (Pulsera, etc..) but the internal layout is clearly different.

Battery voltage is good.  Filament glows at the tube.  No voltage drop when we try to expose, but hear/feel relay engage.  Anyone have insight or experience with this error?  

Thanks


   RE: Mobile C-Arm Report Post
Wed Oct 22 2014Reply from chris?
avatar placemark

Email me at crisb1955@aol.com I maybe able to help


Report Post
Wed Oct 22 2014Reply from ZFSE
avatar placemark

Thanks for the reply.  The hospital had another system and we swapped the ESU--wasn't the issue.  It was the SEA5 board.  Not sure what the part number is.



ex-CBET
ex-AF
Report Post
Tue Oct 28 2014Reply from ZFSE
avatar placemark

FYI, Philips (Australia) says that one board is not an FRU.  They will only sell us the entire assembly (6 boards, case, other components)...