Ring artifact in all fields.

Dear all,

I have problems with CT Aquilion 64 Slices Toshiba.

please help me in service manual.

If you have it, please share to me!

Thank in advance.

Regards,

Hoai Thuong


   RE: CT Scanner Report Post
Sun Nov 23 2014Reply from BMET-KY
avatar placemark

First I would clean the Mylar ring and the detector cover to make sure there is no contrast or other objects on it. Then run your air cals make sure nothing it in the field. You should be able to clean the detector cover without opening the gantry. Just lift the Mylar ring out of the way. If that doesn't take care of it you may have a bad QV2 or ADC card in the detector. If that's the case you can use the DCA program in engineering to determine the channel that's causing the ring