Change date on Alpha

Dead battery caused system to revert back to 1984. GE Logic Alpha Version 1.2. I believe I need hyper terminal to fix. Can anyone provide the solution?


   RE: Ultrasound Systems Report Post
Tue Nov 25 2014Reply from BMET-KY
avatar placemark

I had i similar problem on a Logiq 400 that had a bad timekeeper ram/battery. Here is the procedure that worked for me when I replaced the battery. It may be similar

You will need a cross cable (25 pin to 9 pin serial with a null modem on the 25 pin side.)

The only parameter change in hyperterminal was the baud rate to 9600

At the VTSH> prompt type: (it is case sensitive)

cd MSTPRG
ld BRST.LO
st


You will be prompted for the year (last 2 digits),month (0-12), day (0-31), hour (0-23), minute (0-59).
Then the system serial number.

This loads the entered information into the timekeeper ram.


 