Mon Dec 15 2014

cofigeration error F01 and F11

I have 4008S machine gives me these errors. i checked the upper control module ( all the boards) on an other machine and they worked ok. also i changed the distribution board and still the same. also changed the cables and the HDF module and still the same....

What can i do else ???

please advice.


   RE: Hemodialysis Unit Report Post
Wed Dec 17 2014Reply from strigou
avatar placemark

Hello there;

Unfortunately, I have little experience in Fresenius, but the error codes are these:

 

F01 Configuration

– The HPU DIP switch (CPU2,array 2,DIP 4)has been set (erroneously), the HPU does not answer.

– Correct DIP switch settings.

F11 Configuration

– The HPU answered via the CAN bus,although the HPU DIP switch has not been set.

– Correct DIP switch settings.

 

Your problem could even be corrosion to PCB, due to moisture...

Good luck, and let me know the solution, once you find it.


Report Post
Sun Dec 21 2014Reply from luay
avatar placemark

Dear Strigou,

Thank you for replying my question, i appretiate this.

I tried with this DIP switch , if i put it on the message appears, F01. and when it's off it disappear but the machine failed to Pass the test. 

I tried the control box complete on working machine to decide whether the error from the PCB's or the hydraulic , it worked correctly. so I have something in the communication or in the hydraulic. i will try both and if i find the problem i will tell you.

regards.

Luay


Report Post
Sun Jan 31 2016Reply from Davitik
avatar placemark

Hello. I heve Fresenius 4008s and the problem F11 configeration. What could it be???

Help me please.


Report Post