Charger failed

Arrow sequence stops at 17. Arrow 18 never shows up and skips to 19. Then "Charger Failed" error follows soon after.

I have replaced batteries, charger board, generator driver board, filament driver board and power cap module.

All power levels are there, interlock completes up until failure. I need some ideas, please.


   RE: Mobile C-Arm Report Post
Tue Dec 16 2014Reply from Arthur
Arthur

Hi,

I had a very similar problem and it ended up being the HV Supply Regulator PCB.


Report Post
Tue Dec 16 2014Reply from EJaco
avatar placemark

Thank you Arthur!

I will have the HV Supply Regulator PCB tested as you suggested and the x-ray controller as well while we are at it. I hope either one turns out to be the problem.

Best regards,

EJ