Sun Dec 21 2014

acuson antares dont boot after siemens logo

acuson antares dont boot after siemens logo


   RE: Ultrasound Systems Report Post
Sun Jan 25 2015Reply from ELBAGA
ELBAGA

can you check the io module  leds from 1 to 9

and tell me

elbaga85@hotmail.com



BR

A.ELBAGA
Report Post
Thu Feb 19 2015Reply from greenmedic
avatar placemark

dear sir,

 

i checked io module leds first one glows orange, second didn't glow & all others are glow as green.how can i find out  the problem.

 

Raj venugopal

greenmedichealthcare@gmail.com


Report Post
Thu Feb 19 2015Reply from Jim Carr
Jim Carr

LED 2 is for the hard drive. No light usually means the hard drive has failed. Check the connections first, then replace it. 


Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
 

Report Post
Thu Feb 19 2015Reply from greenmedic
avatar placemark

dear sir

 

i am all reade chaning hard drive but sys is warm up all time


Report Post
Mon Feb 23 2015Reply from Danelis
avatar placemark

the problem probably lies with the RM Pcb. Try removing, cleaning and reseating all the memory and replace the bios battery just in case. If this doesnt work you will need to change the RM pcb.


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.