M326 x-ray generator not available

Can anyone tell me the cause/fix for this error?


   RE: Mobile C-Arm Report Post
Thu Jan 08 2015Reply from chris?
avatar placemark

Does this error code show up on boot up or when you try to fluoro email me crisb1955@aol.com


Report Post
Thu Jan 08 2015Reply from bkm3062
avatar placemark

Hello Chris,

I was a little premature in my request for help. I hadn’t actually started troubleshooting yet, but now I have a little more information.

The error occurs at boot. There’s also a message displayed on the monitor which states “Error in surgery controller”. And when I press the “I” button on the c-arm, in addition to the M326 Error id displays error M050 System Initialization problem.

Looking at the circuit boards, all of the +/- 15v LED’s are lit. All of the circuit boards seem to have all of the correct LED’s lit except for SEA2 Rotor Board.

A1 – A4 are NOT lit. And on the SCPU module, there’s a red LED labeled “STAT” that is flashing. So it looks like I have a stator power problem.

I checked the 220vac coming into the HF Generator assembly and it is ok. I checked the fuses beneath the card rack of the HF Generator and the fuses on the rotor board. All are ok. I have 15v on TP 2 and TP 3. I suspect the rotor board but I can’t be certain.

What are your thoughts?


Report Post
Sun May 17 2015Reply from abogatko
avatar placemark

Quote bkm3062

Hello Chris,

I was a little premature in my request for help. I hadn’t actually started troubleshooting yet, but now I have a little more information.

The error occurs at boot. There’s also a message displayed on the monitor which states “Error in surgery controller”. And when I press the “I” button on the c-arm, in addition to the M326 Error id displays error M050 System Initialization problem.

Looking at the circuit boards, all of the +/- 15v LED’s are lit. All of the circuit boards seem to have all of the correct LED’s lit except for SEA2 Rotor Board.

A1 – A4 are NOT lit. And on the SCPU module, there’s a red LED labeled “STAT” that is flashing. So it looks like I have a stator power problem.

I checked the 220vac coming into the HF Generator assembly and it is ok. I checked the fuses beneath the card rack of the HF Generator and the fuses on the rotor board. All are ok. I have 15v on TP 2 and TP 3. I suspect the rotor board but I can’t be certain.

What are your thoughts?

 

 


Report Post
Mon May 18 2015Reply from abogatko
avatar placemark

The Stat means status of BIST (built in self test). 

Green  Bist OK

Red    Bist Failed

Red Blinking  Bist in progress

The led's on sea2 a1-a4 will not come on till you do x-ray saying anode is turning and up to speed.

Hope this helps

Andrew