Color artifacts

Hello, I am getting a lot of color artifact even at lower gains. It happens on all transducers. The system is a D cart has. I have reseated the front end and it seems better but still not what it should be. Any advice would be great thanks


   RE: Ultrasound Systems Report Post
Mon Jan 26 2015Reply from NGreen
NGreen

This could be a NAIM board or a channel board issue.

If you email me I can set you up with a Hero kit with no restocking fees!!!

Ngreen@acevisioninc.com


NG
Report Post
Mon Jan 26 2015Reply from Kyle_Summit
Kyle_Summit

Color noise/artifacting is typically a NAIM just like Ngreen said.

I would also check the condition of the AFP and Cables leading to is and if you have the time swap channel board 0,1 with 2,3 as pairs and see if it moves the artifact or changes its shape. 

Do you have a photo of the noise you could post?



kyle@mysummitimaging.com

Manager of Global Education & Training

Website: http://www.mysummitimaging.com/

Phone:  1-866-586-3744

24/7
Report Post
Wed Feb 04 2015Reply from BMET-KY
avatar placemark

Thanks for the help, It turned out to be a bad channel board.


Report Post
Mon Feb 23 2015Reply from Trainflyer
avatar placemark

One good way to troubleshoot a bad channel board is to put the system into M-mode (slow sweep helps) and slowly adjust the color gain. Particularly with an intermittent channel noise issue, you may have to wait and watch, but you will see vertical color spikes to at least send you to the channel board area.