I need user manual for Toshiba Ultrasound SSA 340/350. My problem is, the system fails to calculate LMP and EDD; it gives LMP and EDD 0 for whatever BPD value. Please help.
Wed Feb 11 2015
user manual
Wed Feb 11 2015Reply from Sijo
r u getting the LMP and EDD measurements for the HC, AC , FL readings. If u r getting for the rest of the measurements the biometry table for the BPD values are missing. If u can enter it manually do it or else reload the software.
Report Post
Mon Feb 16 2015Reply from Mtalam
Thanks for the response. I don't get LMP and EDD measurements for HC, AC, FL, BPD readings. LMP and EDD values are always set to 0. The report generated is ok except for LMP and EDD zero entries. The system clock has to be setup every time you switch on the machine.
Please help.
Mtalam