U325 Batteries Error 32

I understand that when batteries are bad on U325 board, unit need re-calibration. Does anyone have procedure for this?


   RE: Mobile C-Arm Report Post
Mon Feb 23 2015Reply from sennheiser
avatar placemark

Hi .

Calibration is not required. However, you need to enter the menu confirm the battery. Of course there is also the service password you want. Do not do that even if the system is working.


Report Post
Tue Feb 24 2015Reply from ISE
avatar placemark

I am not sure I understand. Can you give me a step by step please?


Report Post
Fri Feb 27 2015Reply from sennheiser
avatar placemark

Hi .

1 - U325 card is changed in the 2 batteries.

2 - To introduce the battery needs to enter on the service password. There password?


Report Post
Wed Apr 15 2015Reply from jay672
avatar placemark

Is there a step by step procedure for this?


Report Post
Wed Apr 15 2015Reply from ISE
avatar placemark

What I have learned is that after replacing the two batteries on the U325 board, you will still get an error until you can log in with a laptop and special cable to confirm a new battery installation. Password is the II size. So far I have not been able to find someone with that cable. We just keep hitting the prep button to clear the error.


Report Post
Mon Apr 20 2015Reply from ZFSE
avatar placemark

if it's like the 8000, the cable connects to the handswitch socket.  you also need the exposcop software to complete the reset.



ex-CBET
ex-AF
Report Post