CP1000 replenishment pump not pumping sufficient volume

On my CP1000 the replenishment pump motor is running, the belts & gears are fine and don't appear to be slipping, the pumps are turning, the impellers are fine, the vertical pipes are not obstructed, I opened the adjusting screws all the way open.  The tanks are full.  When the replenishment is requested, I can hear the speed increase yet no fluid is pumped up the vertical pipes and out.  I have no replenishment, the system does not consume any chemistry.  All 3 pumps behave the same.  Can anyone tell me how fast the pumps/motor should be turning?  I measured the pump drive gear with a digital tach: 93 RPM normal rate and about 200 RPM in replenishment.  The manual says the speed should double which it does but I must assume I am not reaching the desired speed to move fluid.  The speed is about the same with the tanks removed - so not a bad pump problem.  Does anyone know what voltage should be seen at the motor?  Appreciate any thoughts.


   RE: Film Processor Report Post
Thu Mar 05 2015Reply from ricardsc2
avatar placemark

Dear,

I think that you have a problem with de pump


Report Post
Thu Mar 05 2015Reply from ricardsc2
avatar placemark

I think that you have a problem with the pump,to check this problem can you interchange the pumps


Report Post
Thu Mar 05 2015Reply from ricardsc2
avatar placemark

The CP 1000 is conected to 220 V or 110 V?what kind of voltage do you have in the site?110 V or 220 V


Report Post
Fri Mar 06 2015Reply from elia
avatar placemark

Thanks for the reply.  Fixed now.  All 3 pumps were not pumping - problem was corrosion on the drive shafts slowing down the motor.


Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree