Mon Mar 02 2015

M211

A local doctor's office has a BV Pulsera that gives an M211 error code. Does anyone have a service manual for this c-arm ? I have 2 Philips BV 300 and 5 GEOEC 9800 c-arms at my hospital, but no Pulsera units.


   RE: Mobile C-Arm Report Post
Mon Mar 02 2015Reply from lao
avatar placemark

M211 system not available. The BIST of the timing control board has failed because safety is active. check hand or foot switches. check cabling


Report Post
Mon Mar 02 2015Reply from chris?
avatar placemark

I can help email me crisb1955@aol.com about your problem


Report Post
Mon Mar 02 2015Reply from Chris Brown
avatar placemark

The system is shooting x-rays fine. The M211 error says it is the Camera Iris Postioning Error.

They do not use a foot switch with this unit. The tech stated it made a noise similar to a fan going out before giving the M211 error.

Rebooting the system does not make a difference. Error comes back after power on.


Cbrown1
Report Post
Mon Mar 02 2015Reply from Chris Brown
avatar placemark

Sorry guys, I forgot to say thank you for all your help.


Cbrown1
Report Post
Thu Feb 02 2017Reply from antolin cansigno
avatar placemark
what is the meanning for error M200 iris camera positioning error, PHILIPD BV 300

Report Post