IE 33 with flickering touch screens

I have an E cart with 6.0 software that intermittently has flickering touch screens. I have reseated cables and replaced the right touch screen w/o effect!


   RE: Ultrasound Systems Report Post
Wed May 13 2015Reply from Kyle_Summit
Kyle_Summit

This could be signs of failing ground connection or possibly a bad video card, i would check connections on the video card and potentially swap it with the other card on the UMB, typically when you swap it solves the problem since only one port on the card fails at a time. the main display may stop working though and this would mean you would need a new video card. 

 

feel free to call our technical support team any time, or check out our website for instructions to swap cards. 



kyle@mysummitimaging.com

Manager of Global Education & Training

Website: http://www.mysummitimaging.com/

Phone:  1-866-586-3744

24/7
Report Post
Wed May 13 2015Reply from Bassam
avatar placemark

Check the brightness on the back of the LCD, I've seen that happen on two systems when the brightness is low the screen filckers, the other thing is that replacing the screen didn't help and the other screen is ok, if this doesn't work i'd chek the power in the back of the screens by swapping the power supplies between the screens and see if the issue follows the power source, if so you I'd replace the whole assembly

Good luck


Report Post
Thu May 14 2015Reply from chris k
avatar placemark

thanks for the suggestions i will dig in further


Report Post
Thu May 14 2015Reply from Jim Carr
Jim Carr

Some more things to consider: The backlight inverter control is on the CPM. The brightness pot connects through the Control Panel Connector board. The video for the touch screens goes from one video card thru the PCI bus to the UAVIO, then to the CPM thru the CP Connector board.

If swapping the video cards does not change the symptom, measure the resistance of the grounding wire in the CP to UAVIO cable. I suggest checking the +5V on the CPM, with an oscilloscope if possible.  If all that is OK, the UAVIO, CPM and CP Connector boards are possible causes. AUE provides Hero Kits (and technical support) to our customers for situations like this.


Jim Carr
Director of Service & International Operations
AUS
Toll Free (USA, Puerto Rico and Canada): 866-620-2831
USA office: +1 918-628-2831
Email: Service@advancedultrasound.com and Sales@advancedultrasound.com
 

Report Post
Fri May 15 2015Reply from chris k
avatar placemark

i swapped the control panel and the harness from the UI to the front end we shall see...

 

chris


Report Post
Fri Nov 03 2023Reply from livath14_7015
avatar placemark
A deep clean of the machine and the  change of the  thermal paste on the 
graphics card and motherboard 
processors might reduce this 
issue.

Report Post