Could you explain these error codes?
4028 and 5006
Thanks,
Joe DeJohn
Castle Medical Solutions
Could you explain these error codes?
4028 and 5006
Thanks,
Joe DeJohn
Castle Medical Solutions
check memoskope
check memoskope acqusition cables D1 X10 27
CHECK CONNECT?ON CABLE MEMOSKOPE
Joe -
4028 ---> is usually a bad Memoskop, but occasionally can be a bad connection on the SG cable connection at X-10 (the main interface cable connecting to the c-ram)
Error E4028 Memoskop unexpected acq end signal
Possible trouble: Host registered unexpected acq end signal of the Memoskop.
Possible help: Check faults on the Memoskop acquisition cable (D1.X10.27).
Check Memoskop.
Check connection cables to the Memoskop.
5006 ---> is an indication that S2 is set to "off" on the D-1 PC Board. (this can happen accidentally during troubleshooting, or during R/R of the d-1 board, etc.)
Otherwise, it may be a "glitch", if it does not recur.
Error E5006 Gen. tried X-Ray on without SS relais
Possible trouble: Generator attempts radiation with SS relay switched off.
Possible help: D1, switch S2 in wrong position.
D1 defective.
Good Luck, Wayne