Thu Jun 04 2015

we are having philips ie 33

after switching on the on umb i am getting AA and on the display "no signal power saving mode digital" plz give me the rply


   RE: Ultrasound Systems Report Post
Thu Jun 04 2015Reply from BMET-KY
avatar placemark

Try swapping the two video cards. They are a common failure on these units. If the right one is bad you wont get a display on the main LCD. If the left is bad the touch screens wont display properly. If after swapping and only the touch screens are affected order a replacement video card. Philips only sells the UMB assembly to customers but there are third party vendors who will sell the card.


Report Post
Thu Jun 04 2015Reply from Conquest Imaging - Matt T.
Conquest Imaging - Matt T.

BMET-KY is on the right track. We stock all versions of these cards.



Matt Tomory

VP Sales, Marketing and Training

Conquest Imaging

866-900-9404
Report Post
Fri Jun 05 2015Reply from darla prasad
avatar placemark

sir if we change any board in other system it will ask any password


Report Post
Fri Jun 05 2015Reply from Danelis
avatar placemark

If you replace or swop the Graphics card you will not need a password. When ordering a new graphics card make sure you get the correct type. There are 3 different versions. To check which type you have switch on the system and immediately you will see the type of graphics card shown on the monitor.All the major thrid party ultrasound service companies in the USA will be able to offer you prices on exchange graphics cards.

 

Regards

 

John Williams


Report Post
Fri Jun 05 2015Reply from BMET-KY
avatar placemark

John is correct you shouldnt need a password. Also you dont have to swap with another system. You can just swap the two cards on your system. and see if the problem swaps from main LCD to touch screens. If you have another system with the same type of video cards you could swap both of them and if problem follows order a new video card. I have also had the motherboard or LCD be the problem. Try swapping either of those if the video cards dont change you symptoms. Just make sure its the same cart level. You can also call Conquest Imaging, Summit Imaging, Midwest Imaging, AUE Tulsa, or Ultrasound Gurus they can all offer tech support (free usually) and parts


Report Post
Sat Jun 06 2015Reply from darla prasad
avatar placemark

I have tryed  swapping the two video cards i am getting 608 diagnostic code 


Report Post
Sat Jun 06 2015Reply from darla prasad
avatar placemark

i am getting 608 diagnostic code error


Report Post
Wed Jun 10 2015Reply from Kyle_Summit
Kyle_Summit

Error 608 can be a software error, typically you would reload software if you have that capability. 

were you able to get ahold of a new Video Card just incase one is causing the issues ? or did you only swap left ot right in the umb ?

i look forward to helping call or email anytime!



kyle@mysummitimaging.com

Manager of Global Education & Training

Website: http://www.mysummitimaging.com/

Phone:  1-866-586-3744

24/7
Report Post