Siremobil Compact ERR

   was working fine. now gives error 2080.

now boots with this error and when pushing pedal getting a dark circle with no image


   RE: Mobile C-Arm Report Post
Wed Jun 10 2015Reply from outinleft
avatar placemark

SSW says: Error E2080 Host - powerfail signal active during bootup

Possible trouble: The powerfail signal was active during bootup. 

Possible help: (a) Check powerfail signal wire from power supply M14

(b) Power supply M14 defective

(c) Board D1 defective.

If you need parts, we have all Compact parts in-stock.

- Wayne


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Wed Oct 07 2015Reply from ktm
avatar placemark

thanks Wayne,

after pushing pedal- now get an image, but after 3 shots again a dark circle.

when booting it never goes past the 2 circles half black / half white. My other system completes the cycle and goes into screen with data only.


Report Post
Wed Oct 07 2015Reply from outinleft
avatar placemark

too slow to do this via text. Call me for tech support/parts. Wayne Horsman, Columbia Imaging  (410) 730-4023


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com
Report Post
Wed Oct 07 2015Reply from outinleft
avatar placemark

Any error codes?

Sounds like the memoskop is not booting all the way.

- Wayne