Red Light Alarm

I've had the filter replaced in my EverFlo Q, but the red light still comes on after a few seconds of operation, and there is a steady alarm.  How do I make that stop?  Is there a reset method?


   RE: Oxygen Concentrator Report Post
Mon Jul 27 2015Reply from Len
Len

First - check that the flowmeter is turned up - often it has been turned off.

Chec the oxygen percentage - too low will cause alarm.

Chec the hours run - Thomas compressors in this unit have a life span of 12 - 16000 hours.


Report Post
Sun Dec 04 2016Reply from fjpr325
avatar placemark

I´ve remplaced the 4 ways valves, but the red led is turn on. When I put the CPAP, the sieves start opening,  like oysters...

I don´t kwon, what happen, because, the remplace parts number it´s same that the original remplace, but the body is not similar. The original is black, the remplace is white


Report Post
Sun Dec 04 2016Reply from Len
Len

There was a new supplier for the valve and I think this one was white. They are interchangeable.

When refitting the valve make sure that there is a washer under the screw that goes above the compressor inlet so that the screw does not penetrate a channel in the valve.


Report Post
Tue Mar 26 2019Reply from Shagawa designs
avatar placemark

Where can I buy parts for Sverdlovsk concentrator?


Report Post