Burdick Quest 3.22

After completeing the Touchscreen Calibration successfully and choosing complete, and reboot the screen does not react to touch, but does during calibration. Anyone have a clue ? 


   RE: Electrocardiograph (EKG / ECG) Report Post
Tue Sep 01 2015Reply from TSS Jim
avatar placemark

There are 2 possible answers. If the database gets full the touch screen tends to act unstable. The calibration screen is a separate program than choosing the touchscreen as the input device. With Quest running go into system settings and choose the touchscreen.


Report Post
Tue Sep 01 2015Reply from Doc Fixit
avatar placemark

I have also seen it act that way if the incorrect "monitor" is selected.  Go into the system settings and be sure the correct LCD or CRT is chosen.


Report Post
Mon Sep 14 2015Reply from Einstein
avatar placemark

Easy fix ! you must have a Mouse plugged in in order for the system Mother Board to accept the Ps2 Keyboard. This is also dependant on the SW ver and LCD / CRT setting using PCMCIA Software LCD setup card. 


Report Post
Mon Sep 14 2015Reply from Einstein
avatar placemark

Is there a PCMCIA SW Card for older Ver 2.22 ?   I have the one for 3.22 but does not seen to work with older ver


Report Post