Fri Sep 11 2015

c cff 006 004

I need degassage transducer calibration value


   RE: Hemodialysis Unit Report Post
Fri Sep 11 2015Reply from ChinhNguyen
avatar placemark

Hi Kadri!

Why did you start a new thread here with the same problem?

Did you have your AK95s service manual? There are specific instructions for you to calibrate your transducer at page 100 or chapter 6: Calibration -6.11 degassing pressure transducer calibration.

If you dont have service manual, give me your email and i will send it for you. Hope it helps.


Report Post
Fri Sep 11 2015Reply from KADRI
avatar placemark

I have not calibrator


Report Post