Mon Sep 21 2015

what causes of any transvaginal probe did not work at any socket although any convex probe at any socket is work perfectly ?

when i connect the transvaginal probe at  any sockat (sonoace 3200 medison ultrasound) is not work perfectly..and i replaced this transvaginal probe with anew transvaginal probe AND SAME PROBLEM IS FOUND.

AND ANY CONVEX PROBE IS WORK PERFECTLY

 

PLEASE HELP ME ..THANK YOU VERY MUCH


   RE: Ultrasound Systems Report Post
Mon Sep 21 2015Reply from Estrella
Estrella

Hi, gabr

try to do a backup restore, maybe the calibration of the transvaginal probe is wrong!



Best regards,
Estrella.


If you have some technical problem with your imaging equipment, send me an email, and i will reply asap to help you.

Email: jmstreya@gmail.com
Report Post
Mon Sep 21 2015Reply from Bassam
avatar placemark

I have few questions for you:

Are these probes new to the system? Have you used them on the system before? If so, what changed since the last time you used them? If not, check to make sure the system had the proper license to use the probe, also are the calculation packages for the probe available to choose

This should give you a place to start

Thank you


Report Post