leica cm 1850 temp problem

my unit was working fine ..maintaining -23 C but then it stopped cooling more than -12 C. I REFILLED THE GAS BUT NO CHANGE. SAME PROBLEM.

I REPLABCED THE COMPRESSOR WITH NEW ONE .STILL IT IS AT -12 C..

 

CAN ANYONE HELP ME ..HOW TO GO MORE LOW TEMP.WHAT CAN BE THE ISSUE.

 

Arun


   RE: Cryostat Report Post
Mon Oct 05 2015Reply from Ross
avatar placemark

 

Did you replace the filter/dryer? Could it be plugged? Is the defrost solenoid leaking hot gas into the evap? What are the high and low pressures?


Report Post
Tue Oct 06 2015Reply from ARUNJARYAL
ARUNJARYAL

adjusted the amount of refreigeant to be filled.

i think i filled more than what was needed. released some amount and i got -17 c.

but i m thinking to replace drier also.

No ..defrost valve not leaking gas otherwise -17 c not possible.


Report Post
Thu Oct 08 2015Reply from MedWrench MM
MedWrench MM

Try cleaning the coils on the condenser cooling, looking inside the compressor area for coolant leaks and have someone certified with ac compressors check the high and low pressures and the coolant levels.

 

Answer provided by Paul M via LinkedIn.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Oct 20 2015Reply from ARUNJARYAL
ARUNJARYAL

cleaned coils also..but adjusted gas volume approriate level for max temp . got -22 c now.


Report Post
Tue Oct 20 2015Reply from Ross
avatar placemark

 

You should set it to a lower temp, to test it. Check to see that it will get down to -30C, or lower.


Report Post
Fri Oct 23 2015Reply from ARUNJARYAL
ARUNJARYAL

NO..ITS MAX GOING UP TO -22 C. NEED TO CHANGE EXPANSION SLOT AND DRYER...THEN WE CAN REACH UPTO -30 C MAY BE


Report Post
Fri Oct 23 2015Reply from Ross
avatar placemark

The filter-dryer should normally be replaced during a compressor replacement. A quick check to help to determine if it is partially plugged is to see if the downstream tubing attached to the filter-dryer is cold (or cooler than the tubing attached to the upstream side). If there is any cooling there, then the filter-dryer is acting as an additional expansion device, and there is less cooling capacity available in the chamber.


Report Post
Sat Oct 24 2015Reply from Hawkeye Biomedical LLC
Hawkeye Biomedical LLC

EV valve is the problem.





 
Report Post
Mon Oct 26 2015Reply from ARUNJARYAL
ARUNJARYAL

Yes ...absolutely right.


Report Post
Wed Dec 02 2015Reply from Rudydog
avatar placemark

Had the same issue. There was an obstruction in the line causing the issue. Someone had installed a flip valve type nozzle that had fell apart internally limiting the R404 freon flow. Our failed suddenly too.......as the valve fell apart inside. Don't believe this valve was original equipment. Just throwing out a possible scenario.


Report Post