connector artifact. can I secure or is this a pcb replacement


   RE: Ultrasound Systems Report Post
Tue Oct 06 2015Reply from BMET-KY
avatar placemark

If you know how to solder you can repalce the connector. Here is info from and earlier thread.

"Amphenol part # 97-3102A-14S-6S. It doesnt come with the correct rotation, but you can pop the retaining ring off the back side and rotate it to the correct position then reinsert the retaining ring. I ordered from Newark part # 93F1165 they were $13 last time I needed them"


Report Post
Mon Oct 19 2015Reply from Dennis Balow
avatar placemark

Thanks for your response. When I disassembled the connector, I found a couple female pins to be pretty loose. I re secured them. Hooked up to chicken heart in the shop, worked great. Staff had further trouble with artifact when in practical use. Ordering a new FEC for unit. I'm going to hold on to the FEC I have now for troubleshooting/parts.