Recently our C-arm only consistently only showed center view of the X-ray image with outer circle (about 50% diameter) of view all blocked. No error message given from theC-arm.

Anyone has any idea what's going on? Thanks.


   RE: Mobile C-Arm Report Post
Tue Nov 17 2015Reply from www.PI.solutions
www.PI.solutions

Are the collimator blades in the way?



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post
Tue Nov 17 2015Reply from MChang
avatar placemark

Yes, I believe collimator blades are in the way, which showed themselves around the border of the limited X-ray view.

Any idea what's wrong? How to resolve this problem?


Report Post
Wed Nov 18 2015Reply from www.PI.solutions
www.PI.solutions

Take off the tube cover and see if you are able to manually move the blades. If they are really easy to move then the motor is bad.



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post
Wed Nov 18 2015Reply from MChang
avatar placemark

Yes, they are easy to move and spring works fine. What will take to replace this motor? Is Philips Service the best to do this repair?


Report Post
Wed Nov 18 2015Reply from www.PI.solutions
www.PI.solutions

Philips used to have an FCO for this issue but not sure if they still do. They may charge you for a new collimator. May be worth looking for a motor, or a used collimator. I could do the calibrations. Where are you located? Call Block Imaging. A few years back I used to change the motors for Philips but I think I remember them stopping giving out the motors.



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post
Thu Nov 19 2015Reply from MedWrench MM
MedWrench MM

You can always reach out to our sponsor Philips Multi-Vendor Services.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Thu Nov 19 2015Reply from MChang
avatar placemark

We are in Seattle area.

I'll start to look for collimator motor.

Just a thought: if you just remove the spring, which returns collimator blades to close, can you keep those blads to maintain open?


Report Post
Tue Nov 24 2015Reply from www.PI.solutions
www.PI.solutions

Im sure you could but you do not want to do that. FDA frowns on that :)



Jamison Ashby, MBA

Owner: Precision Imaging solutions

www.PI.solutions
Report Post