Tue Dec 22 2015

DEFECTIVE DATA MESSAGE

When trying to run a test the EKG starts up for a few seconds and then says -NO ANALYSIS ATTEMPTED DUE TO DEFECTIVE DATA.   Before the diagnosis it says  " -DEDECTIVE ECG-"    We have been using this machine with no problems prior to this issue.  The leads were then checked to make sure they were properly placed and there didn't seem to be any issues.  We ran another test and it came out the same.  

Any ideas on what is wrong?


   RE: Electrocardiograph (EKG / ECG) Report Post
Tue Jan 26 2016Reply from LL
avatar placemark

This happens when there is a faulty connection between the patient and the machine.  Suspect the patient cable or the cable socket.  It may be as simple as the csable not being plugged in well.


Report Post