Wed Jan 13 2016

Output Port expected output for HomeFill II?

I have a Platinum XL set up to work with the Homefill II compressor. I set the oxygen concentrator at 2.5-3LPM and verfied that the Oxygen output from the main port is >93%. 

I connected the compressor through the rear output port and it has been unable to fill the bottle due to low oxygen input from the concentrator. I tested the output from the rear output port from the concentrator and it is indeed greater than 90%, however the flow output is less than 1LPM. 

Should the output port flow be so low? I cheated the set-up to make sure that the compressor worked by connecting it to the front port of the concentrator instead of the correct port, and it did manage to get high O2 and fill the bottle intermittently. However, this is explicitly contraindicated by the manufacturer. 


   RE: Oxygen Concentrator Report Post
Wed Jan 13 2016Reply from medicalbridges
avatar placemark

UPDATE: I revisited the Ventura II Manual. THe necessary input LPM for the compressor needs to be greater than 2LPM, hence its inability to operate with the Platinum XL at hand. 

Any insight into increasing the output port flow would be greatly appreciated. 


Report Post