I am in a remote area of Guatmala. Our otoscope Welch Allyn 74710 will not light. I checked our power source- good. When turned on the indicater light does not come on and neither head will light. As there is no Circuit breaker or fuse. What is a likely possibility? It is the only scope we have and if a field repair is possible I would really appreciate the guidence.
Don S
Actually there should be a couple of 1/8 amp fuses on the board. RPI Parts in California carries replacement power boards for the 74710, What they do carry is at http://www.rpiparts.com/pdf/WelchAllyn0914.pdf
The indicator light may simply be a burned out neon bulb. But check the microswitches on the wand holders, it will only light up one at a time. If both are off-hook, neither one lights up.
Other than that I don't know what to tell you. I haven't had one in my hands in a few years. You might want to drop the folks at RIP an Email and see if they can help you out.
Report Post
Thanks Ray. They only sell to suppliers and repair. I will email them and see if the can help me. The part pic they had shows the fuses, but the unit I have has only two transformers a capacitor and the switch.
Report Post
Hey Don S. Could you solve your problem. If not, give me a call. I live here in Guatemala and I´ll be happy to help to fix your problem. I have experience with medical equipment and I happen to have fixed one of this scopes. I have electronics engineering studies. Cellphone: 5019-9246