Diagnostic No 0002

>   Hello
>  How are you doing doing ?
> Place can you help me solve this problem on PHILIPS HDI 5000.
> SPM 2000 Signal Processing model 7500-1952-06A
>
> Diagnostic No 0002
>
> Philips.PMS SN.02J929

 


   RE: Ultrasound Systems Report Post
Tue Apr 05 2016Reply from dferguson
dferguson

0002 diagnostic error is a front end issue; however, it could also be caused by a transducer.  It is a power draw issue.  I would remove all probes first and try each one individually to see if it one of them.  If not that, then possible AIM+ or Channel Board.


 
 

Report Post
Thu May 19 2016Reply from Jason T
avatar placemark

Hello,

I have 7 probes, 2 working and 5 that come up with this same error message 002. You seem to know your way around this machine pretty well. Other than the 5 probes needing repair is there any other reason 2 only may be working. The 2 that work are L12-5 and C8-5, also the 2 that end in -5. Does the 5 mean anyting, ie if it is a current draw issue do teh 2 working draw less current or some other significance. Could just be a coincidence. thank you for your assistance in advance. 