errors

our centrifuge has an error -4- on display panel please help. when i open their is no physically damage ,burned, cold, components/cable....I have tried mains reset several times,what might be the problem


   RE: Centrifuge Report Post
Thu Apr 07 2016Reply from MedWrench MM
MedWrench MM

If you need immediate assistance you can always reach out directly to our Centrifuge sponsor Ozark Biomedical.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Apr 08 2016Reply from amorales
avatar placemark

I have the service manual, send me your email and I will send it to you.


Report Post
Thu Aug 25 2016Reply from Amscopolaris
Amscopolaris
– 4 – Communication Error Communication error between control board and power board Error consequence Drive switch off . No brake effective Error cause • Loose contact in flat ribbon cable • Control board (A3) defective • Power board (A1) defective Error code reset MAINS RESET after standstill. If the MAINS RESET performed before standstill,