our centrifuge has an error -4- on display panel please help. when i open their is no physically damage ,burned, cold, components/cable....I have tried mains reset several times,what might be the problem
Thu Apr 07 2016
errors
Thu Apr 07 2016Reply from MedWrench MM
If you need immediate assistance you can always reach out directly to our Centrifuge sponsor Ozark Biomedical.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Apr 08 2016Reply from amorales
Thu Aug 25 2016Reply from Amscopolaris
– 4 – Communication
Error Communication error between control board and power board
Error
consequence
Drive switch off . No brake effective
Error cause • Loose contact in flat ribbon cable
• Control board (A3) defective
• Power board (A1) defective
Error code reset MAINS RESET after standstill.
If the MAINS RESET performed before standstill,
Report Post