Need info ,What is the normal life span of these devices? Newlife elite?
Mon May 02 2016
No beep? also light code info please?
Mon May 02 2016Reply from Brian Hincy
Mon May 02 2016Reply from mjaholwell
If I'm understanding your question correctly, it appears as though your power interrupt alarm is failing to sound. To correct this, try replacing the 9V battery that powers it. If that fails to correct the fault, check both the buzzer and the cabling. Let me know if this helps. Cheers!
Report Post
Mon May 02 2016Reply from mjaholwell
As a follow-up to your second question, the life of service of this device is as long as you can have you can have it meet OEM specifications while it's still supported. The bacteria filter needs to be replaced every 10,000 hours and the compressor (Thomas 2560/2660) needs to be rebuilt every 20,000 hours.
Report Post