fault code 052 _ Versapulse Holmium Yag 80/100

Good Morning everybody

I'have a old Versapuse Holmium Yag PowerSuit

and randomly appear the fault code 052 consequently the touch screen freeze. Someone know which cause this error code ??  

thank you in advance

Best Regrdas

Gabriele


   RE: Surgical Laser Report Post
Wed Sep 14 2016Reply from ChristineMae
ChristineMae

Low voltage power supply is not supplying to proper voltage you need. double check the connections on the LVPS. If no change I suggest replacing the LVPS.

 

hope this helps.




Christine Ricoy

C.L.R.T.

Universal Medical Lasers
Report Post
Thu Dec 08 2016Reply from Sab31
avatar placemark

I have the same intermittant error on my machine: Versapulse 100w powersuite

After speaking to Lumenis UK

052 error code

Fault: Main Screen panel failed to respond in designated time to the computer

Remedy: Replace Display

Hope this helps


Report Post
Fri May 13 2022Reply from Bakhtiyarr
avatar placemark

Hi, I have VersaPulse PowerSuite Holmium 20 Watt laser. F304 (energy low) appears all the time. What is the issue? 


Report Post
Fri May 13 2022Reply from Wayne-UML
Wayne-UML

Energy low means that the internal energy detector is registering that the energy produced is less than 80% of what you are requesting. It may simply require a recalibration, but it also may indicated an optical issue.


Wayne Delgado
Senior Laser Engineer
www.umlasers.com