Good Morning everybody
I'have a old Versapuse Holmium Yag PowerSuit
and randomly appear the fault code 052 consequently the touch screen freeze. Someone know which cause this error code ??
thank you in advance
Best Regrdas
Gabriele
Good Morning everybody
I'have a old Versapuse Holmium Yag PowerSuit
and randomly appear the fault code 052 consequently the touch screen freeze. Someone know which cause this error code ??
thank you in advance
Best Regrdas
Gabriele
Low voltage power supply is not supplying to proper voltage you need. double check the connections on the LVPS. If no change I suggest replacing the LVPS.
hope this helps.
I have the same intermittant error on my machine: Versapulse 100w powersuite
After speaking to Lumenis UK
052 error code
Fault: Main Screen panel failed to respond in designated time to the computer
Remedy: Replace Display
Hope this helps
Hi, I have VersaPulse PowerSuite Holmium 20 Watt laser. F304 (energy low) appears all the time. What is the issue?
Energy low means that the internal energy detector is registering that the energy produced is less than 80% of what you are requesting. It may simply require a recalibration, but it also may indicated an optical issue.