error 4021
I need service manual with error code
since thanks
Usually the memoskop needs to be replaced. email us for a quotation at wayne@columbiaimaging.com
Below is the error text from the Siemens service software:
Error E4021 Memoskop does not answer
Possible trouble: Host repeats a task 3 times, error E4021 is then entered.
Possible help: If the Memoskop C / CE 100 / C-MOD / C-SUB / C-SUB &MOD was replaced, try "Erase Harddisk".
(Memoskop, Technical setup, menu "Diagnostics" - "Erase Harddisk")
Note: Ask your customer before you start erasing !
All images will be lost !
Check for additional error messages E2006 and E4024.
Possibly Memoskop defective.