Thu Jul 28 2016

error 4021

error 4021

I need service manual with error code

since thanks


   RE: Mobile C-Arm Report Post
Thu Jul 28 2016Reply from outinleft
avatar placemark

Usually the memoskop needs to be replaced. email us for a quotation at wayne@columbiaimaging.com

 

Below is the error text from the Siemens service software:

 

Error E4021 Memoskop does not answer

 

Possible trouble: Host repeats a task 3 times, error E4021 is then entered. 

 

Possible help: If the Memoskop C / CE 100 / C-MOD / C-SUB / C-SUB &MOD was replaced, try "Erase Harddisk". 

(Memoskop, Technical setup, menu "Diagnostics" - "Erase Harddisk")

Note: Ask your customer before you start erasing !

All images will be lost !

Check for additional error messages E2006 and E4024.

Possibly Memoskop defective.


Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com