Fri Aug 05 2016

Volume Factor Calibration

Cannot get this measurement in range. It just stays at 70 +/- 5. Any ideas?


   RE: Ventilator Report Post
Fri Aug 05 2016Reply from MedWrench MM
MedWrench MM

Try reaching out directly to our Ventilator sponsor Soma Technology.

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Aug 05 2016Reply from BiomedSalim
avatar placemark

I'm not looking for somewhere to send it. I AM the biomed. Just looking for fellow biomed tips/tricks/advice. Thanks though...


Report Post
Mon Aug 08 2016Reply from Dean
Dean

Just a couple of questions...

Are you using the RP50 parabolic restrictor?

Have you performed the exhalation valve calibration?

Have you gone on past the volume factor and tested the pressure relief valve adjustment?

Any of these three can have an effect on the volume factor.


Dean Stephens EET, CBET, CRES
Senior Biomedical Equipment Tech
Intermediate Imaging Equipment Tech
(Not my first rodeo...)

Report Post
Mon Sep 26 2016Reply from BiomedSalim
avatar placemark

Yes.

Yes.

Yes, pressure relief valve won't go above 75.


Report Post