I have a Blanketrol II that is causing the line isolation monitor to alarm. Maximum current draw during normal operation is @ 7 amps, heating or cooking. Anybody have any experience with this issue?
Fri Aug 05 2016
Joe Miriani BioMed Tech III
Mon Aug 08 2016Reply from Bobby S.
What is the leakage current. Check your grounds. The motor or compressor may be shorting to ground. See if you can disconnect each one, one at a time and see if the leakage drops down.
Bob Sullivan
Report Post
Mon Aug 08 2016Reply from Bobby S.
What is your leakage current? Check your grounds. Might be the compressor or the fan shorting. Disonnect each on , one at a time and see if it drops.
Bob Sullivan