Fri Aug 19 2016

Propylene glycol

Is adding 2 oz of propylene glycol a MUST in order to keep the micro temp LT, MODEL 749 disinfected?

What is the purpose of adding propylene glycol?

 


   RE: Patient, Forced-Air Report Post
Fri Aug 19 2016Reply from MedWrench MM
MedWrench MM

What specific equipment are you referring to? Manufacturer/Model?


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Aug 19 2016Reply from mskehan14
avatar placemark

Cincinnati Sub-Zero Micro-Temp LT Model 749


Report Post
Fri Aug 19 2016Reply from Triad1313
avatar placemark

The MicroTemp LT model 749 does not require Propylene Glycol it was a manufacturers suggestion.

The Decontamination procedure is to add 27 ml of household beach to the full reservoir run the device for five minutes then drain.  Fill the device three times running it for five minutes each time then draining.  After no bleach is detected refill with 873 ml of distilled/sterile distilled water.  This should be repaeted quarterly at a minimum.