Hi,

 I own a Fluke prosim 8, while trying to automate the controls i've a challenge in stop data for SpO2 for say 10 Seconds  and bring it up after a while for performing this action is there any controls to either stop flushing SpO2 data or to put prosim to sleep and bring it up after a duration? 


   RE: Analyzer Measurement & Simulator Report Post
Wed Aug 24 2016Reply from MedWrench MM
MedWrench MM

Hi Pradz,

I spoke with my contact at Fluke concerning your question. Here was his repsonse:

"There is no way to put the ProSim 8 asleep when using the ProSim 8 with the SpO2 Test Module as a stand-alone SpO2 functional tester.   I have also taken a look at the Ansur Automation Software and there is also no provision to set a delay before the ProSim 8 starts the SpO2 “simulation”. 

It is a new feature that could be investigated the next time the Ansur Software and/or ProSim 8 firmware is upgraded."

 

You can speak with John at Fluke directly by contacting him at john.shirlock@fluke.com


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Thu Aug 25 2016Reply from Pradz
avatar placemark

Hi Jonathan,

    Thank u for the Quick reply! 

Regards,

Pradz


Report Post
Thu Aug 25 2016Reply from MedWrench MM
MedWrench MM

You are most welcome, Pradz.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Wed Aug 31 2016Reply from medicus
avatar placemark

Do you only use the ansur as interface?
e.g if you use Labview for automation you can easyls create delays bevor measurement


Report Post
Wed Sep 07 2016Reply from Pradz
avatar placemark

Hi,

We are just sending commands via COM port from a java framework.

Right now we are using the self test command from prosim to ensure this data inactivity.