I have change new battery so overcome the error F56 but the problem still exist which let whole system down..
please advise how I could go about solving this Error F56 on the console/screen.
Desperately need your help!
I have change new battery so overcome the error F56 but the problem still exist which let whole system down..
please advise how I could go about solving this Error F56 on the console/screen.
Desperately need your help!
You need to go into the service mode and check the positive and negative voltage numbers with the voltage of the battery packs on the left side of the unit on the charge board. If there off that can cause your problem. Did you check the voltage of each battery before you put them in? Battery voltage can vary and if you put too many high battery in the positive pack can cause problem. If you need the service manual let me know. Is this a Mux 100 series or a Mux 200 series unit.
Thanks for the info given, I will try and go through all the batties one by one and confirm. If problem still exsist i will adivise.
I have a mux 100 series X-ray machine.
thanks alot.
Benjamin Ipato
I need a service manual too, if can be able to provide would be greatly appreciated.
the machine is mux 100 series.
please provide me a diagram of how to go about installing the batteries if possible.
Thanks
Benjamin Ipato
Quote Benjamin Ipato
I need a service manual too, if can be able to provide would be greatly appreciated.
the machine is mux 100 series.
please provide me a diagram of how to go about installing the batteries if possible.
Thanks
Benjamin Ipato
I have a similar problem. The battery volatge on the positive and negative sides is similar. I have adjusted the recorded system voltage to closely match the battery voltages. However, the problem recurs on full charging and now on startup. How do I solve this error?