I have an error 4025. In the morning the techs state the system freezes with the split screens on each monitor. This is during boot up. They will power off and then back on, system comes up normal.
Wed Aug 31 2016
error 4025
Wed Aug 31 2016Reply from MedWrench MM
If you are in need of immediate assistance please feel free to reach out directly to our Mobile C-Arm sponsors. They are always happy to help out when they can.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Wed Aug 31 2016Reply from outinleft
All 4000 series errors are Memoskop Errors, but I don't see 4205 in the list.
If you can swap out the Memoskop with another unit (to verify that the memoskop is the problem) we can repair/replace the memsokop for you.
Otherwise, you are left with replacing the memoskop & hoping it fixes your problem.
What is the part # or model # and DOM of your Memoskop?
- Wayne Horsman wayne@columbiaimaging.com