Error code 0091. Does anyone know what this error code means. Our book stops at 0090. All PS voltages are within spec. Thanks in advance, Serge
Wed Sep 14 2016
HDI 5000 error 0091
Thu Sep 15 2016Reply from mgoldy
Fri Sep 16 2016Reply from Serge.N
Fri Sep 16 2016Reply from mgoldy
Fri Sep 16 2016Reply from MedWrench MM
If you are in need of immediate assistance please feel free to reach out directly to our Ultrasound System sponsors. They are always happy to help out when they can.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Fri Sep 16 2016Reply from Mauro
It could also be the DVS or APM Module as it recognizes its hard drives as components of the HDI-5000.
Press Superkey + FR (frame Rate) to expand the ERROR Screen and a second time for it to disapear.
Report Post
Thu Sep 29 2016Reply from bxanchezo
Indeed.
Error 091 itisn't related to a change of hardware, that's error 08 (Actual and Expected configuration do not match)!!