Tue Oct 04 2016

Imaging Noisy

The equipment Start good but after few minutes start getting Distortion in the Transducer imaging area


   RE: Ultrasound Systems Report Post
Sun Oct 09 2016Reply from MartinJade
avatar placemark

You didn't mention any power levels, so I will go with 'standard' settings. Do you have more than one probe to try in the same port to see if the symptoms duplicate? Transducers that have been subjected to rapid deceleration (aka dropped) can build up heat quickly...if the symptoms repeat with a number of probes, then I would look towards the transducer interface...if it's one or two, but others work fine, the have the transducers analyzed.


Report Post
Mon Oct 10 2016Reply from parksron61
avatar placemark

A known issue with Zonares is what they call the headlight artifact. If this is what you are experiencing you will see an artifact in 2 areas. One at the side and the other in the middle of the image. We have had that on several units and they had to go to Zonare to resolve.


Report Post
Sun Aug 20 2017Reply from kecman
avatar placemark

Did the Zonare solved the problem?


Report Post
Sat Mar 21 2020Reply from GMT
avatar placemark

Hi, How did you solve the headlight artifact problem?


Report Post
Tue Jun 22 2021Reply from abir
avatar placemark

Comment on peut résolu se problème des artefacts du phase svp.

 


Report Post