Wed Oct 19 2016

Display Blank

Hi Folks,

 I've got a MAC 5500 with a blank display. I shone a flashlight across the screen and I could make out text on the screen. I swapped the display with a new one out of the box but the blank display issue continued. I tried the new display on another machine and there were no issues.  Could the main board be the problem? It's usually a cable issue.

Thanks in advance.


   RE: Electrocardiograph (EKG / ECG) Report Post
Thu Oct 20 2016Reply from biomed683
avatar placemark

check Video Flex cabel. may be it is faulty.. and do confirm that the network cabel is sitted well on the connector..


Report Post
Thu Oct 20 2016Reply from RJBiomed
avatar placemark

Thanks Jamil

I tried the original display with another MAC and it worked without issues. I'm leaning towards not a display or cable issue. I haven't come across a main board causing this before but it could happen.


Report Post
Thu Oct 20 2016Reply from S. Rooks
avatar placemark

I read your efforts and can truely see why you may eliminate the display assembly as a possible cause, but keep in mind that the brown ribbon cable (2034923-001 or -002) traces can fatigue and partially seperate causing it to "black out" intemittently. By moving the ribbon you may have relieved the stress point and now it appears functional because the traces are touching once again.

To be sure its not in the main board, as you commented, try installing the display from your other machine on this unit. I would be willing to bet it will work just fine.

The inverter ribbon is an inexpensive part to replace (and should be a spare part on hand for any Biomed dealing with MAC's) compared to a main board or display assembly and has almost always been the cure for me.  

Please feel free to call for further tech support, it's free and i'm happy to trroubleshoot it with you.

Steve Rooks, BMET

Southeastern Biomedical Associates

1-888-310-7322


Biomedical Equipment Technician with Southeastern Biomedical 1-888-310-7322
Report Post
Thu Oct 20 2016Reply from jimsdavis
avatar placemark

always the flat ribbon cable the provides the backlite voltage, changed them many times. be sure to get the one that matches you machine as there are a few variations


Report Post
Thu Oct 20 2016Reply from AJ_Diaz
avatar placemark

I have 15 of these EKGs and 9 times out of 10, it has been that brown/orange display flex cable.

We had an instance where we replaced the backlight itself, but that was the exception to the rule.


Report Post
Fri Nov 11 2016Reply from RJBiomed
avatar placemark

Thanks again for the assistance. It ended up being the main board. Once replaced and reloaded with software the display came up ok.


Report Post
Tue Dec 20 2016Reply from SIU2014
avatar placemark

i'd have to agree with everyone else on this 9 times out 10 its the flex display cable


Report Post