Will detect maternal but will not pick up fetal signals. Have used different test equipment and cables to verify no equipment error and equip self test shows no errors.
Thu Nov 03 2016
GE 259CX-B / 250CX-SERIES FETAL , MONITOR
Thu Nov 03 2016Reply from MedWrench MM
If you are in need of immediate assistance I would reach out directly to Pacific Medical. They may be able to help.
Just tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Thu Nov 03 2016Reply from Greg Smith
There are different pcb's that control maternal ecg and fetal ecg in this monitor. There is a strong possibility that the FECG board is bad. The part number for the FECG board is 2025177-008. Please call for assistance with this part I will be glad to help you.