TEMPERATURE

My autoclave is heating up to about 260 degrees, but not holding temperature. what could be the problem?


   RE: Sterilizers/Autoclaves Report Post
Fri Nov 04 2016Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Sterilizers/Autoclaves sponsor RPI. They are very knowledgeable and always happy to help.

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Sun Nov 06 2016Reply from Sterilizer Tech 3
avatar placemark
Constant bleed valve may be blocked or not enough flow going throug it. Check steam traps that they are getting rid of condensation. And your steam source is enough to keep a dynamic pressure that stays consistent

Report Post
Sun Nov 06 2016Reply from arthur sanchez cbet,cet,cfcc
avatar placemark
Replace. Door gasket,open up unit and run a cycle.whith case open,look for any steam leak.